How to calibrate the sensor when always in environment of 1000 ppm?

0 Likes
1 Solution
AditiB_81
Moderator
Moderator

Hi,

In your case, the CO2 concentration is not changing. For such cases, you can use the FCS scheme (Forced Compensation Scheme) which lets you calibrate the sensor to a fixed concentration of CO2. Yes, the upper limit for this scheme is set to 900 ppm but I'm following up on this request with the internal team and we need some more time. I'll revert back to you on this asap.

Thanks

Aditi

View solution in original post

0 Likes
2 Replies
DaTra_4942216
New Contributor II

If possible you may need to think of using FCS scheme time to time. However at this point of time, the FCS is capped at 900 ppm unfortunately. So, even if you try to input 1000 ppm, it will cap it to 900 ppm. In the future if the upper cap can be increased to bigger value I will update with new information

0 Likes
AditiB_81
Moderator
Moderator

Hi,

In your case, the CO2 concentration is not changing. For such cases, you can use the FCS scheme (Forced Compensation Scheme) which lets you calibrate the sensor to a fixed concentration of CO2. Yes, the upper limit for this scheme is set to 900 ppm but I'm following up on this request with the internal team and we need some more time. I'll revert back to you on this asap.